Topic 8 min read

Effect of the staircase on current sampling.

Latest updated: June 2, 2023

Cyclic voltammetry is one of the most commonly used electrochemical techniques. The purpose of this technique is to stress the cell by applying a voltage scan and measure the resulting current ITOT, which is actually the sum of the capacitive IC  and faradaic currents IF.

 

ITOT= IF + IC

 

Because of the potentiostat’s digital architecture, the voltage scan, dE/dt, is performed by applying a small voltage step, dE (Fig. 1).

 

Fig. 1: Voltage scan. Green: Linear. Blue: staircase.

 

An introduction to digitally sampled current is given in the article “Cyclic Voltammetry – how to obtain good results”.

 

Depending on where, and how, the sampling is managed, you will obtain more faradaic or more capacitive current.

 

NOTE: For good voltage control resolution it is important to mimic the perfect ramp (as close as possible to analog behavior).

 

So let’s compare analog & discrete scanning.

 

As discussed below, an electrochemist’s objective is to perform the most accurate ramp, corresponding to an analog ramp. So it is critical to understand how the current is sampled.

 

 

In which condition is my digital sampling equivalent to analog sampling?

 

NOTE: in CV sim, the model is based on analog behavior. See article “How to measure reaction kinetics parameters using cyclic voltametry

 

A priori there is no need to obtain equivalence between analog and digital voltammetry at a unique sampling value. There are numerous reviews dealing with this subject [1-4] and more recently this comparison has been extensively investigated by Montella in a series of four papers. [5-8]

 

 

Is one point enough?

It has been demonstrated that according to the species under study, the sampling at one point, in a range between 30% to 50% of the step duration, is equivalent to a true linear (analog) scan [5] (Fig. 2). Intrinsically, it is difficult to optimize the settings as this parameter has to be known before the measurements are carried out.

 

Fig. 2: Voltage scan. Sample at 30% of step duration, dt. Sampling indicated by the black dot.

 

 

Is over the step averaging helpful?

Montella investigates the subject in another way. If averaging is performed on the whole step [8], it has been demonstrated that the sampling yields nearly the same current as the true linear voltage scan. This has been shown on reversible and quasi-reversible electrochemical reactions without ohmic drop compensation and double layer effect.

 

Fig. 3: Voltage scan. Sample over 100% of the step duration, dt. Sampling is shown by the black line.

 

Focus in EC-Lab: This parameter is managed by the following setting.

 

 

To obtain true analog behavior, use the Analog Ramp Generator (ARG) add on. This option allows also fast sampling, up to 1 MV/s as the voltage scan definition is not limited by the voltage control resolution.

 

 

[1] A.J. Bard, L.R. Faulkner, Electrochemical methods, Fundamentals and Applications, second ed., John Wiley & Sons, Inc., New York, 2001.

[2] R.G. Compton, C. Batchelor-McAuley, E.J.F. Dickinson, Understanding Voltammetry: Problems and Solutions, Imperial College Press, London, 2012.

[3] R.G. Compton, E. Laborda, K.R. Ward, Understanding Voltammetry: Simulation of Electrode Processes, Imperial College Press, London, 2014.

[4] J.-M. Savéant, Elements of molecular and biomolecular electrochemistry, An Electrochemical approach to Electron Transfer Chemistry, John Wiley & Sons, Inc., Hoboken, NJ, 2006.

[5]: C. Montella , Further investigation of the equivalence of staircase and linear scan voltammograms. I- Sampling conditions for reversible reactions involving soluble species. JEAC, 796 (2017) 96-107.

[6]: C. Montella , Further investigation of the equivalence of staircase and linear scan voltammograms. II – Effects of electron transfer kinetics, Ohmic drop and double-layer charging. JEAC, 799 (2017) 194-205.

[7]: C. Montella , Further investigation of the equivalence of staircase and linear scan voltammograms. III-Averaged-current staircase voltammetry applied to electrochemical reactions involving adsorbed species. JEAC, 808 (2018) 348-361.

[8]: C. Montella , Further investigation of the equivalence of staircase and linear scan voltammograms. IV – Averaged-current staircase voltammetry applied to electrochemical reactions involving soluble species. JEAC, 817 (2018) 73-88.

 

 

 